    DPS 2 Pump Motor

    1 have a 250 G2HO, model E250LHAGF, S/N 05503246, the connector between the DPS Module & the DPS Pump motor got hot enough to melt the 2 pieces together. I checked the current & the pump motor is pulling 27 amps. Does anyone know what the current draw should be for this? I have a service manual, but it does not specify the current rating.
    TIA
    Jim

  2. Member
    Join Date
    Sep 2021
    Location
    Salem, wi
    Posts
    1,034
    #2
    Highly suggest you also post here https://www.etecownersgroup.com/categories

    there are several G2 dealers that post on that site. They may have more direct experience on your question.

    what will extremely helpful is a copy of the diagnostics report. I would expect a fault code to be triggered .
    Last edited by Bighare2; 05-12-2023 at 07:18 AM.

  3. Member my492vs's Avatar
    Join Date
    Apr 2005
    Location
    Van Alstyne Texas
    Posts
    122
    #3
    Thanks Bighare2, I will try the elec owners group. It did generate the open circuit fault code (211) when the plug melted. I replaced the connector & that alarm cleared. It is not generating the overcurrent alarm (216), but the wires are getting hot when the pump is running. I have not taken it to a dealer, so I do not have a diagnostic report.